Re: Post Sort Quenching

BIGOS@Darwin.Stanford.EDU
Mon, 10 Feb 1997 08:22:28 +0000 (PDST)

Dave Parks and I spent some time looking at loss of
fluorescence of PE to try to figure out why low power
Fascans did better than high power jet-in-air sorters at
measuring PE stained cells. We determined the PE
saturation curve and found that the dye saturated
significatly at 100mW. Moreover, we observed that with
several hundred milliwatts of 488nm illumination, PE was
kicked into some unusal state from which it didn't
recover. Re-analysis of sorted cells minutes later had
lost up to 45% of their origional fluorescence. With
less laser illumination, this effect was minimized. For
FITC we did not notice much change - 5% at most. We did
not look at Cy5PE at that time.

>From our experience we would expect that with the low
laser power that you are using (15mW) the losses for
FITC and PE will be small.

-Marty Bigos
Stanford Shared FACS Facility


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu